I am trying to convert the dental CBCT scan into a 3d model to be printed. however when I do that the end result does not look realistic and the jaw looks elongated and the compressed.
When I upload the cbct it show a warning " multi-frame image. If slice orientation or spacing is non-uniform, then the image may be displayed incorrectly" however, I do not think this is possible. Is there is a way to fix this?
I have the Dicom files uploaded if anyone is interested in checking it.
I encountered another problem when I try to using dicom files from medical ct scans. The render is not smooth and look like layers as seen in the picture on the following link
It seems that the volume has spacing between slices that is about 10x larger than pixel spacing within a slice. Use Crop Volume module to resample the volume with isotropic voxel size by checking Isotropic spacing checkbox and selecting B-spline interpolation. After this segment the cropped/resample volume using Segment Editor module.