I was today using CBCT data exported as dcm multiframe images of patients.
I used 3D Slicer to import the .dcm file and when it was importing it gave a warning in advance mode saying " Multi-frame image. If slice orientation or spaicing is non-uniform then the image may be displayed incorretly. Use with caution"
And after importing anyway when i use volume rendering to visualize the bones, i realized the rendered image is scaled or compressed or skewed i dont know what exactly but “not proper”.

So i went back and used gdcmconv to uncompressed the .dcm file and used the uncompressed dcm file in the 3D slicer. Then the volume rendering was accurate.

Why is this ? Is there a way of properly importing the compressed cbct data directly to 3D Slicer without using gdcmconv and cli ?

I cannot share the screenshots due to i did not seek permission from the patient to use the data in public.

Thanks

This information is not enough to even guess what could have gone wrong. Follow instructions in the DICOM FAQ and if it doesn’t solve the problems then let us know.

Thank you for the quick reply Prof Iassoan. I will go through it and let you know and come back with more details.

And just now i figure out there is a x3 time scaling added when i uncompressed the dcm with gdcmconv command when i measured the distances.

Is there a way to scale it back or something is horrible wrong from the onset ?

Thank you Prof @lassoan.

The DICOM Patcher module corrected the problem including the scaling problem i had.

Also just out of interest i tried the volume --> spacing and when i gave same spacing value again i think the problem settled because image spacing in volume information section was 0.3, 0.3 and 1.0 mm so when i changed everything to 0.3 mm it again seems it was working.

Thank you

1 Like

@manjula - I’m glad you were very attentive to the spacing issues in the data and were able to find a workable solution. To help make this process more robust, please scan a phantom with known dimensions using the scanner in question and share the data for testing.

I guess it’s a Dolphin CBCT scanner. Developers of this device thought that slice thickness tag could be used to define image geometry, which is of course incorrect.

This is the rule in DICOM patcher that fixes it:

@pieper and @lassoan ,

It is a i-CAT KAVO dental CBCT. I will try to scan a phantom with that but it is not my scanner. It belong to colleague of mine and he send me this for a test run with 3D Slicer for a project we are planning for. I am not sure but i will try.

thanks

Sounds good. Perhaps you can explain to your colleague that the scanner needs to be calibrated/tested to ensure accurate measurements. In any case best of luck.

1 Like

Does the image come directly from the scanner or it has been processed by some software? Maybe the problem is not in the Dolphin software (in previous sample images that I got, Dolphin was in the manufacturer tag, but maybe it was just used for processing).

If it is straight from the scanner then the scanner manufacturer should be made aware of their probable non-conformance with the DICOM standard.

If you can provide a sample scan of a phantom then we can give you more specific information about which required DICOM tags are missing from their images.

1 Like

I will tell him next week when i meet him.

Is the metdata is of any use ? because It is already available !

Yes, metadata is useful, could you copy-paste it here? (make sure you don’t remove any tags just replace values that contain personal information by ***).

1 Like

I have pasted two metdata sets.

Metdata Set 01

[0008,0005] SpecificCharacterSet ISO_IR 100 CS 10
[0008,0008] ImageType [3] ORIGINAL, PRIMARY, AXIAL CS 22
[0008,0012] InstanceCreationDate 20170629 DA 8
[0008,0013] InstanceCreationTime 082742 TM 6
[0008,0014] InstanceCreatorUID 99999 UI 6
[0008,0016] SOPClassUID 1.2.840.10008.5.1.4.1.1.2 UI 26
[0008,0018] SOPInstanceUID 2.16.840.114421.80462.9552036653.9615108653 UI 44
[0008,0020] StudyDate 20170629 DA 8
[0008,0021] SeriesDate 20170629 DA 8
[0008,0022] AcquisitionDate 20170629 DA 8
[0008,0023] ContentDate 20170629 DA 8
[0008,0030] StudyTime 082742 TM 6
[0008,0031] SeriesTime 082742 TM 6
[0008,0032] AcquisitionTime 082742 TM 6
[0008,0033] ContentTime 082742 TM 6
[0008,0050] AccessionNumber SH 0
[0008,0060] Modality CT CS 2
[0008,0070] Manufacturer Imaging Sciences International LO 30
[0008,0080] InstitutionName Imaging Sciences International LO 30
[0008,0081] InstitutionAddress Hatfield, PA ST 12
[0008,0090] ReferringPhysicianName PN 0
[0008,1010] StationName RONTGEN SH 8
[0008,103e] SeriesDescription LO 0
[0008,1050] PerformingPhysicianName PN 0
[0008,1070] OperatorsName rtg PN 4
[0008,1090] ManufacturerModelName 17-19 LO 6
[0009,0100] Unknown Tag & Data 9267 LO 4
[0009,0101] Unknown Tag & Data 1 IS 2
[0010,0010] PatientName ************* PN 16
[0010,0020] PatientID ********** LO 10
[0010,0030] PatientBirthDate ******** DA 8
[0010,0040] PatientSex * CS 2
[0010,2160] EthnicGroup SH 0
[0010,4000] PatientComments LT 0
[0018,0022] ScanOptions PORTRAIT CS 8
[0018,0050] SliceThickness 0.300 DS 6
[0018,0060] KVP 120 DS 4
[0018,0090] DataCollectionDiameter 170.00 DS 6
[0018,1000] DeviceSerialNumber ICU-080462 LO 10
[0018,1020] SoftwareVersions 1.9.3.13 LO 8
[0018,1150] ExposureTime 4 IS 2
[0018,1151] XRayTubeCurrent 5 IS 2
[0018,5100] PatientPosition CS 0
[0019,1000] Unknown Tag & Data ISI_ACQU_1 LO 10
[0019,1043] Unknown Tag & Data 1 IS 2
[0019,1050] Unknown Tag & Data 64.3059095263 DS 14
[0019,1051] Unknown Tag & Data 73.0000000000 DS 14
[0019,2000] Unknown Tag & Data Pan EFOV DAP
[0019,2001] Unknown Tag & Data 9deSGSZl+9BZr9jdo0hvGbDe+I4 LO 28
[0019,2002] Unknown Tag & Data 4.586000^0.138000^0.138000 ST 50
[0020,000d] StudyInstanceUID 2.16.840.114421.80462.9552036462 UI 32
[0020,000e] SeriesInstanceUID 2.16.840.114421.80462.9552036477.9583572477 UI 44
[0020,0010] StudyID SH 0
[0020,0011] SeriesNumber 0 IS 2
[0020,0012] AcquisitionNumber 0 IS 2
[0020,0013] InstanceNumber 1 IS 2
[0020,0032] ImagePositionPatient [3] -384.000, -384.000, 100.000 DS 26
[0020,0037] ImageOrientationPatient [6] 1.000000, 0.000000, 0.000000, 0.000000, 1.000000, 0.000000 DS 54
[0020,0052] FrameOfReferenceUID 2.16.840.114421.80462.9552036477.9583572477 UI 44
[0020,1040] PositionReferenceIndicator LO 0
[0020,1041] SliceLocation 100.000 DS 8
[0021,1032] Unknown Tag & Data 18.537371 DS 10
[0021,1034] Unknown Tag & Data 8.883619 DS 8
[0028,0002] SamplesPerPixel 1 US 2
[0028,0004] PhotometricInterpretation MONOCHROME2 CS 12
[0028,0008] NumberOfFrames 576 IS 4
[0028,0010] Rows 768 US 2
[0028,0011] Columns 768 US 2
[0028,0030] PixelSpacing [2] 0.300, 0.300 DS 12
[0028,0100] BitsAllocated 16 US 2
[0028,0101] BitsStored 16 US 2
[0028,0102] HighBit 15 US 2
[0028,0103] PixelRepresentation 1 US 2
[0028,1050] WindowCenter 1100 DS 4
[0028,1051] WindowWidth 4500 DS 4
[0028,1052] RescaleIntercept 0 DS 2
[0028,1053] RescaleSlope 1 DS 2
[0032,4000] RETIRED_StudyComments LT 0
[7fe0,0010] PixelData OB 0

MetData set 02

[0008,0005] SpecificCharacterSet ISO_IR 100 CS 10
[0008,0008] ImageType [3] ORIGINAL, PRIMARY, AXIAL CS 22
[0008,0012] InstanceCreationDate 20170914 DA 8
[0008,0013] InstanceCreationTime 092126 TM 6
[0008,0014] InstanceCreatorUID 99999 UI 6
[0008,0016] SOPClassUID 1.2.840.10008.5.1.4.1.1.2 UI 26
[0008,0018] SOPInstanceUID 2.16.840.114421.80462.9558692675.9621764675 UI 44
[0008,0020] StudyDate 20170914 DA 8
[0008,0021] SeriesDate 20170914 DA 8
[0008,0022] AcquisitionDate 20170914 DA 8
[0008,0023] ContentDate 20170914 DA 8
[0008,0030] StudyTime 092126 TM 6
[0008,0031] SeriesTime 092126 TM 6
[0008,0032] AcquisitionTime 092126 TM 6
[0008,0033] ContentTime 092126 TM 6
[0008,0050] AccessionNumber SH 0
[0008,0060] Modality CT CS 2
[0008,0070] Manufacturer Imaging Sciences International LO 30
[0008,0080] InstitutionName Imaging Sciences International LO 30
[0008,0081] InstitutionAddress Hatfield, PA ST 12
[0008,0090] ReferringPhysicianName PN 0
[0008,1010] StationName RONTGEN SH 8
[0008,103e] SeriesDescription LO 0
[0008,1050] PerformingPhysicianName PN 0
[0008,1070] OperatorsName rtg PN 4
[0008,1090] ManufacturerModelName 17-19 LO 6
[0009,0100] Unknown Tag & Data 8809 LO 4
[0009,0101] Unknown Tag & Data 1 IS 2
[0010,0010] PatientName ******** PN 12
[0010,0020] PatientID ******** LO 10
[0010,0030] PatientBirthDate ******** DA 8
[0010,0040] PatientSex * CS 2
[0010,2160] EthnicGroup SH 0
[0010,4000] PatientComments LT 0
[0018,0022] ScanOptions PORTRAIT CS 8
[0018,0050] SliceThickness 0.300 DS 6
[0018,0060] KVP 120 DS 4
[0018,0090] DataCollectionDiameter 170.00 DS 6
[0018,1000] DeviceSerialNumber ICU-080462 LO 10
[0018,1020] SoftwareVersions 1.9.3.13 LO 8
[0018,1150] ExposureTime 4 IS 2
[0018,1151] XRayTubeCurrent 5 IS 2
[0018,5100] PatientPosition CS 0
[0019,1000] Unknown Tag & Data ISI_ACQU_1 LO 10
[0019,1043] Unknown Tag & Data 1 IS 2
[0019,1050] Unknown Tag & Data 77.6498551190 DS 14
[0019,1051] Unknown Tag & Data 124.0000000000 DS 14
[0019,2000] Unknown Tag & Data Pan EFOV DAP
[0019,2001] Unknown Tag & Data o6fVI9wQoGd8/4ckK5VByBD58Sg LO 28
[0019,2002] Unknown Tag & Data 4.586000^0.138000^0.138000^0.138000 ST 58
[0020,000d] StudyInstanceUID 2.16.840.114421.80462.9558692486 UI 32
[0020,000e] SeriesInstanceUID 2.16.840.114421.80462.9558692501.9590228501 UI 44
[0020,0010] StudyID SH 0
[0020,0011] SeriesNumber 0 IS 2
[0020,0012] AcquisitionNumber 0 IS 2
[0020,0013] InstanceNumber 1 IS 2
[0020,0032] ImagePositionPatient [3] -384.000, -384.000, 100.000 DS 26
[0020,0037] ImageOrientationPatient [6] 1.000000, 0.000000, 0.000000, 0.000000, 1.000000, 0.000000 DS 54
[0020,0052] FrameOfReferenceUID 2.16.840.114421.80462.9558692501.9590228501 UI 44
[0020,1040] PositionReferenceIndicator LO 0
[0020,1041] SliceLocation 100.000 DS 8
[0021,1032] Unknown Tag & Data 18.537371 DS 10
[0021,1034] Unknown Tag & Data 8.883619 DS 8
[0028,0002] SamplesPerPixel 1 US 2
[0028,0004] PhotometricInterpretation MONOCHROME2 CS 12
[0028,0008] NumberOfFrames 576 IS 4
[0028,0010] Rows 768 US 2
[0028,0011] Columns 768 US 2
[0028,0030] PixelSpacing [2] 0.300, 0.300 DS 12
[0028,0100] BitsAllocated 16 US 2
[0028,0101] BitsStored 16 US 2
[0028,0102] HighBit 15 US 2
[0028,0103] PixelRepresentation 1 US 2
[0028,1050] WindowCenter 1100 DS 4
[0028,1051] WindowWidth 4500 DS 4
[0028,1052] RescaleIntercept 0 DS 2
[0028,1053] RescaleSlope 1 DS 2
[0032,4000] RETIRED_StudyComments LT 0
[7fe0,0010] PixelData OB 0

@lassoan @pieper

Jsut out of curiosity, did you manage to obtain any useful information from the metdata ? or do I have to see the possibility of a phantom object scan ?

Sorry for the slow response, both @pieper and I are quite busy at RSNA. I’ve checked the metadata and the problem is the following:

Slice position must be specified for each slice. However in the ISI i-CAT device, only one position is specified using [0020,0032] ImagePositionPatient and probably the developers thought that by adding a [0018,0050] SliceThickness tag will specify slice positions within the image. That is incorrect. Instead, values that are the same for all frames (image orientation, pixel spacing, etc.) should be specified in tag (5200,9229) and values that are different in each slice (such as fields position) should be specified in tag (5200,9230). For example:

(5200,9229) SQ (Sequence with undefined length #=1)     # u/l, 1 SharedFunctionalGroupsSequence
(0020,9116) SQ (Sequence with undefined length #=1)     # u/l, 1 PlaneOrientationSequence
    (0020,0037) DS [1.00000\0.00000\0.00000\0.00000\1.00000\0.00000] #  48, 6 ImageOrientationPatient
(0028,9110) SQ (Sequence with undefined length #=1)     # u/l, 1 PixelMeasuresSequence
    (0018,0050) DS [3.00000]                                #   8, 1 SliceThickness
    (0028,0030) DS [0.597656\0.597656]                      #  18, 2 PixelSpacing
...

(5200,9230) SQ (Sequence with undefined length #=54)    # u/l, 1 PerFrameFunctionalGroupsSequence
(0020,9113) SQ (Sequence with undefined length #=1)     # u/l, 1 PlanePositionSequence
    (0020,0032) DS [-94.7012\-312.701\-806.500]             #  26, 3 ImagePositionPatient
(0020,9113) SQ (Sequence with undefined length #=1)     # u/l, 1 PlanePositionSequence
    (0020,0032) DS [-94.7012\-312.701\-809.500]             #  26, 3 ImagePositionPatient
...
1 Like

Thank you for the information. It has been very useful.

Further, I have noticed that few CBCT’s that i have worked with in the past to import data in to 3D Slicer is having the same problem but did not knew what was wrong at the time but never with the conventional CT’s.

Are CBCT manufacturers are notoriously not compiling to DICOM format ? is this something non compliance is mainly a problem with CBCT manufacturers ?

Yes, CBCT and microCT manufacturers often have DICOM compliance issues. Maybe they have fixed the problems in their current products but not released updates for the earlier models.