It looks quite good. It seems you are either missing some slices (or there are extra slices of the same region that interfere with the reconstruction). To diagnose the issue, please send us all the image slice positions occurring in your data set by follow these steps:
- Open DICOM browser
- Click on the patient
- Check
Advanced
checkbox - Click `Metadata button
- Enter
Image
in the search box - Click
Copy all files metadata
button to copy image position information on the clipboard - Paste it to an empty Excel sheet
- Upload the Excel sheet to OneDrive, Dropbox, or any other file sharing service, and paste the link in your response (taking a screenshot of the sheet and attaching it is not enough, as only a small fraction of the data would be visible)